8. Error Messages
If an error occurs during the measurement, the measurement is
interrupted and an error message, e.g. «
ERR 3
», is displayed.
* Please consult your doctor, if this or any other problem occurs repeatedly.
9. Safety, Care, Accuracy Test and Disposal
Safety and protection
Instrument care
Clean the instrument only with a soft, dry cloth.
Cleaning the cuff
You can machine wash the
cuff cover
at 30°C (do not iron!).
Accuracy test
We recommend this instrument is tested for accuracy every
2 years or after mechanical impact (e.g. being dropped). Please
contact Microlife-Service to arrange the test (see foreword).
Disposal
Error
Description Potential cause and remedy
«
ERR 1
» Signal too
weak
The pulse signals on the cuff are too
weak. Re-position the cuff and repeat the
measurement.*
«
ERR 2
» Error signal
During the measurement, error signals were
detected by the cuff, caused for instance by
movement or muscle tension. Repeat the
measurement, keeping your arm still.
«
ERR 3
» No pressure
in the cuff
An adequate pressure cannot be generated
in the cuff. A leak may have occurred.
Check that cuff and pump ball are correctly
connected and that the cuff is not fitted too
loosely. Replace the batteries if necessary.
Repeat the measurement.
«
ERR 5
» Abnormal
result
The measuring signals are inaccurate and
no result can therefore be displayed. Read
through the checklist for performing reli-
able measurements and then repeat the
measurement.*
«
HI
»
Pulse or cuff
pressure too
high
The pressure in the cuff is too high (over
300 mmHg) OR the pulse is too high (over
200 beats per minute). Relax for 5 minutes
and repeat the measurement.*
«
LO
»
Pulse too low The pulse is too low (less than 40 beats per
minute). Repeat the measurement.*
)
If you think the results are unusual, please read through the
information in «Section 1.» carefully.
•
This instrument may be used only for the purpose described in
this booklet. The manufacturer cannot be held liable for damage
caused by incorrect application.
•
This instrument comprises sensitive components and must be
treated with caution. Observe the storage and operating condi-
tions described in the «Technical Specifications» section!
•
Protect it from:
−
water and moisture
−
extreme temperatures
−
impact and dropping
−
contamination and dust
−
direct sunlight
−
heat and cold
•
The cuffs are sensitive and must be handled with care.
•
Only pump up the cuff once fitted.
•
Do not use the instrument close to strong electromagnetic fields
such as mobile telephones or radio installations.
•
Do not use the instrument if you think it is damaged or notice
anything unusual.
•
Never open the instrument.
•
If the instrument is not going to be used for a prolonged period
the batteries should be removed.
•
Read the additional safety instructions in the individual sections
of this booklet.
Ensure that children do not use the instrument unsuper-
vised; some parts are small enough to be swallowed.
WARNING:
Under no circumstances, however, may you
wash the inner bladder! Always remove the sensitive
bladder from the sleeve before washing and replace it care-
fully again afterwards.
Batteries and electronic instruments must be disposed of in
accordance with the locally applicable regulations, not with
domestic waste.