![background image](/i/uni-t/145436/uni-t-utm-1204a-ut204a/h/uni-t-utm-1204a-ut204a-012.png)
(2)AC voltage ( )
Range
Resolution Accuracy
Overload protection
4.000V
40.00V
400.0V
600V
1mV
10mV
100mV
1V
(1%+5)
(1.2%+5)
600V DC/AC
Input impedance: 10M
//not less than 100pF
Frequency response: 40Hz~400Hz
AC conversion type:
AC conversion is to guarantee consistence with
valid value by sine wave input and correction
reading by way of response mode of average value.
(3) Resistance (
)
Range
400.0
4.000K
40.00K
400.0K
4.000M
40.00M
Resolution
100m
1
10
100
1K
10K
Accuracy
(1.2%+2)
(1%+2)
(1.2%+2)
(1.5%+2)
Overload protection
600Vp
(4) Diode test ( )
Range Resolution
Accuracy
Overload protection
1mV
To display approximate
positive pressure drop.
(Open circuit voltage is
about 1.48V.)
600Vp
(5) Conductance test ( )
Range Resolution
Accuracy
Overload protection
600Vp
Buzzer will ring if it is
less than or equal to
10
. (Open circuit
voltage is about 0.45V.)
0.1
Note: Buzzer will ring or not ring if measured
resistance exceeds 10
during conductance test.
(6) Capacitance ( )
Range
Resolution
Accuracy
Overload protection
600Vp
4nF
40nF
400nF
4
µ
F
40
µ
F
µ
F
0.001nF
0.01nF
0.1nF
0.001
µ
F
0.01
µ
F
0.1
µ
F
(4.0%+3)
(5.0%+10)
To measure under RELATIVE measurement mode;
(7) Frequency (Hz)
Range
10Hz
100Hz
1kHz
10kHz
100kHz
1MHz
10MHz
Resolution
0.001Hz
0.01Hz
0.1Hz
1Hz
10Hz
100Hz
1kHz
(Reading is only for reference.)
Accuracy
Overload protection
(0.5%+3)
600Vp
Sensitivity:
300mV rms if
100kHz;
600mV rms if
100kHz;
800mV rms if
1MHz;
(8) Temperature ( )
Range
-40
~
1,000
Accuracy
Overload protection
-40
~0
0
~400
400
~1,000
(8%+5)
(2.5%+3)
(3.0%+3)
Plug-in resistance
of 1K
Note:
1) There is no voltage protection for temperature
gear. It is not allowed to insert electrified
conductor into jack to avoid burnout of 1K
resistance;
100