2) Spot K type thermocouple (Ni-Cr~Ni-Si) is only 
suitable to temperature measurement of less 
than 230

. Rod type temperature sensor shall 

be used for temperature measurement of more 
than 230

.

(9) DC current (     )

Range

40.00A

600A

0.01A

1A

Accuracy

(2%+5)

Overload protection

600A DC/AC

Note:
Current measurement function must be between 
0

 and 40

. Current direction is from bottom to 

top for positive reading during DC current 
measurement. (As shown in Figure 11, panel is 
on the top and bottom cover is on the bottom.)
Please do not loosen the trigger suddenly after 
pressing. As a sensitive device, Hall element will 
be sensitive to heat and mechanical stress to 
different extents except magnetic sensitivity. 
Collision will cause short-term reading variation.
More correct      measurement can be guaranteed 
by following operation methods:

Press the trigger and open clamp head to 

fetch measured conductor by clamp head then 
loosen trigger slowly until clamp head is closed 
completely. Please check if measured conductor 
is in the middle of clamp head or not. Additional 
reading error of 

1.0% may be caused if failing 

to place it in the middle of clamp head;

To remove clamp head on the current conductor;
To press REL

 key for display resetting;

To repeat step 

;

To gain more correct  reading by above 

    measurement steps;

(9) AC current (     )

Range

40.00A
600A

0.01A
1A

Accuracy

Overload 

protection

600A DC/AC

Resolution

Resolution

(2.5%+8)
(2.5%+5)

Frequency 

response

50Hz~60Hz

Note:
Current measurement function must be fulfilled 
between 0

 and 40

. Frequency response: 

50Hz~60Hz;
Instable or incorrect inductive reading with less 
than 10 words may be displayed in AC current 
gear and it will not influence measurement result.
Please do not loosen trigger suddenly. As a 
sensitive device, Hall element will be sensitive to 
heat and mechanical stress to different extents 
except magnetic sensitivity. Collision will cause 
short-term reading variation.
AC conversion type:
AC conversion is to guarantee consistence with 
valid value by sine wave input and correction 
reading by way of response mode of average 
value.

X. Maintenance(shown in Figure 13)

     Warning: Please remove testing rod before 
     opening bottom cover to avoid electric shock.
1. General maintenance
A. The clamp meter must be repaired and served 
    by qualified professional repair personnel or 
    designated repair department.
B. To clean the shell periodically by dry cloth. 
    However, it is not allowed to use detergent 
    with abrasive or solvent components